tailing still there
Posted: Fri Jul 06, 2012 1:58 am
hi,
according to suggestions you gave to me, i changed the vent filter and split vent line for both back and front . Here is my rinse(DCM:HEXANE=1:1). The problem is still there. by the way ,i scan the chroma, but i don't know how to post or attach it
There is another question: i am usign 5975 GC-MS. I use HS to inject sample(split ratio=50), sample to to one column which analyse c6 to c10, another column is connection MS. The initial tem is 60, aggetate tem is 90. I ran cals std very day. now the problem is i spike 50ul of 200ppm cals to a vial which has 10ml solvent used 20g sodium bisufate dissolve in 4L ioniazed water. I detect c6,c10 and toluene using FID. c10 is always low, sometimes is half response to c6 and toluene. I also spike 15ul of 200ppm to a vial which has 10ml same solvent. The all respond is close. I change gold seal, clean syringe,change liner and column, the problem is still there. Other instruments have the same problem,but it's not stable, sometime is good, sometimes is bad.
could you please give me some suggestions?
thank you so much
rebecca
according to suggestions you gave to me, i changed the vent filter and split vent line for both back and front . Here is my rinse(DCM:HEXANE=1:1). The problem is still there. by the way ,i scan the chroma, but i don't know how to post or attach it
There is another question: i am usign 5975 GC-MS. I use HS to inject sample(split ratio=50), sample to to one column which analyse c6 to c10, another column is connection MS. The initial tem is 60, aggetate tem is 90. I ran cals std very day. now the problem is i spike 50ul of 200ppm cals to a vial which has 10ml solvent used 20g sodium bisufate dissolve in 4L ioniazed water. I detect c6,c10 and toluene using FID. c10 is always low, sometimes is half response to c6 and toluene. I also spike 15ul of 200ppm to a vial which has 10ml same solvent. The all respond is close. I change gold seal, clean syringe,change liner and column, the problem is still there. Other instruments have the same problem,but it's not stable, sometime is good, sometimes is bad.
could you please give me some suggestions?
thank you so much
rebecca