Page 1 of 1

Problem in GC-450

Posted: Thu Jul 05, 2012 9:34 am
by arfeen66
Dear All,
We are using Varian(now bruker) GC-450for Natural gas analysis, there is a prblem in air pressure(EFC11) which is stable at start of test but gradually decrease & results in turned off Flame of FID. i have checked the Leakage but did't find any leakage. please help me to sort out the problem.

Re: Problem in GC-450

Posted: Thu Jul 05, 2012 9:54 am
by Peter Apps
It sounds as if the air supply to the instrument is inadequate - check the pressure at the cylinder, and make sure that the supply line is not obstructed, and that there are not excessive volumes being taken off elsewhere.

Peter

Re: Problem in GC-450

Posted: Thu Jul 05, 2012 10:51 am
by arfeen66
other GC of same brand using the common supply of air shows no problem(300ml/min)

Re: Problem in GC-450

Posted: Thu Jul 05, 2012 11:28 am
by Peter Apps
other GC of same brand using the common supply of air shows no problem(300ml/min)
Probably a faulty EPC then, time for a service call.

Peter

Re: Problem in GC-450

Posted: Fri Jul 06, 2012 10:09 am
by arfeen66
thanks sir for your promt reply