vahur » Fri Jun 15, 2012 7:29 am
Impossible to tell without a *lot* more information:
Column type (the exact brand and name)?
Column dimensions and particle size?
Flow rate (or what was the dead time)?
The exact retention times of your peaks?
The widths of your peaks (or what was the plate count)?
Did you see significant tailing (what was the asymmetry factor)?
Column: Agilent Hypersil C18 4,0x125mm, 5um
Flow rate: 0,8ml/min
asymmetry: (D2, D3) 1,221; 0,866
width: 0,416; 0,4365
dead time: 1,35
tr: 11,43; 12,12
in most cases of RP the 2 compounds will not fully seperate.
we use to have a method using merck lichroshpere that we used and we switched to a cosmosil cholester column.
it performs a lot better for these 2 compounds.
the difference between the 2 is the presence of a double bound in one, and a single bound in the other on one of the positions.
this gives them a difference in molecule "volume", and cholester rigid structure gives a better selectivity for them.
nacalai have an application on line for it on their web-site
D2 has 1 methyle group also more if wikipedia has it right.
Have you tried Acetonitrile?
I haven´t tried acetonitrile. What does it change if I switch between MeCN and MeOH?