I will go for one by one following step to increase resolution.
1)Reduce injection volume.
2)Prepare a new mobile phase and filter it and sonicate it.
3)Reinstall column to check to reduce the dead volume.
4)As KM suggested minimize tubing diameters and lengths.
5)Resolution mainly depends on peak sharpness and difference in retention time. Dirty frit gives broad peak. By reversing the column and cleaning it with either Methanol, Acetonitrile or any solvent which has been used in the previous analysis.
6)Still u are not getting resolution change to new column as some time column loose its efficiency and peak become broad.
6)Try to go for column with amide linkage or any other type of L1 column.
7)If ambient temperature written, you can go up to 30 degree centigrade. Temperature helps to sharpen the peak but at same time decrease retention time.
8)Play with Mobile phase composition as described adjusting USP method. As u have already done by increasing the water content as per USP, some time decreasing the water also make less retention difference but sharpen the peak and increase the resolution.
9)You can have a option for reducing the particle size, diameter of the column.
10) You can try and see to reduce Methanol as per USP criteria, as methanol has more viscosity it may broaden the peak.
As USP can achieve the resolution 9.0, you will definitely reach it. If someone has more ideas please give suggestions.
Thanks,
Nawaz
this is a very nice list of things to do.
1) I will try this, Injection volume is 10 µL, I will try 5 µL first to see if it has any effect.
2) I have done this several times, it was the first thing I did. But definitely good advice.
3) I didn't even consider dead volume being an issue, I did reinstall the column several times and switched it out, but just dumb luck that I tried this one.
4) I have already minimized all the tubing as much as I can.
5) I have actually reversed the column already and flushed with 100% Acetonitrile and 90:10 Acetonitrile:Water. This actually improved my resolution, but not enough, and further backflushing has not improved it at all.
6) I am currently using a Supelco Discovery C18 4.6 mm x 150 mm x 5 µm
7) Good advice, didn't realize I can go to 30C
8) I played with the mobile phase composition the more water and less acetonitrile gave me the best resolution, but I didn't try reducing the water from the 50:25:25 listed in the USP. I can try this.
9) I have a waters symmetry C18 3.9mm x 150 mm x 5 µm, I can try using the smaller diameter column to try to increase the resolution.
10) I may be misreading the USP, I thought it allowed for the change of only one of the minor components in a ternary mixture. If it allows the change of both minor components, I would change the methanol, but I can give the methanol change a try anyway, maybe do a 55:25:20 mobile phase.
Thanks, for this wonderful advice, I will give it a try.