RT advancement in sugars RP HPLC

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

10 posts Page 1 of 1
Hi
I was performing sugar analysis on our Agilent 1260 HPLC. Today I got the sugar peaks with advanced RT. What might be the reason and how can I overcome it? I am worried....

Waiting for a reply....


Dinesh babu.
What might be the reason and how can I overcome it?


We are not mind readers! :roll: If you told us what you were doing (what column packing, what dimensions and flow rate, what mobile phase, what sugars, etc.) someone might actually be able to help.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Yes, sorry.
I was using C18 column as provided by the company for carbohydrate analysis. The flow rate is 1.4ml/min and the mobile phase was 75% ACN in water.

Is there any other details that could be helpful? Please let me know.

Dinesh
Do you pre-mix your mobile phase and run isocratically?
Time flies like an arrow. Fruit flies like a banana.
With these conditions (75% ACN) this must be an amino column?

What is your sample matrix?
A. Carl Sanchez
Yes, sorry.
I was using C18 column as provided by the company for carbohydrate analysis. The flow rate is 1.4ml/min and the mobile phase was 75% ACN in water.

Is there any other details that could be helpful? Please let me know.

Dinesh
Hi
I usually work at a flow rat of 1.0 ml/min. Maybe it can help you.
Byes
Sugars are not retained by reversed-phase mechanism. You might get some minimal retention at 90-95% ACN but this will be attributed to polarity of residual silanols (HILIC). Flow rate does not change void volume or k' prime. You need to use HILIC column (bare silica, aminopropyl, etc.). Also you need to make sure that you have appropriate detection technique and that what ever you see in chromatogram is indeed your sugar or not some artifact of detection
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
i do face the same problem my MP is 79 ACN + 21 water with 2 ml flow rate, RI detector same Agilent 1260 HPLC instrument...

Column is Waters 10 micron amino column with 4.6 X 250 mm...

was not able to sort the problem

thanks a lot for pointing out..

Actually i have raised a complaint to waters when i brought this column.

since our customer was using the same column for more than 4 years their RT was at 2.8 min, whereas in my brand new column i was getting RT at 12-13 min...

so the waters have replaced the column with the brand new one, even in that i was getting RT at 12-13 min..

since the results were fine...the issue has been closed..

now u r link has cleared the mystery...

thanks a lot once again
10 posts Page 1 of 1

Who is online

In total there are 18 users online :: 2 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry