I can't find anything there, like S/N calc, or print extended performance to SST, or insert data to the sequence like multiplayer, dilution... for auto calc etc...
Thanks
Advertisement
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
Hello brasileo,I was working with chemstation, and now in my new job, I use EZchrom.
I can't find anything there, like S/N calc, or print extended performance to SST, or insert data to the sequence like multiplayer, dilution... for auto calc etc...
Thanks
Hello Jackus
Hello brasileo,
we had the same problem when we moved from CHS to EZCh. Agilent local representative made a macro for us which has functionality as S/N calculation in Chemstation. Everything is described in Powerpoint presentation (including necessary files).
Have a nice day.
http://rapidshare.com/files/453960098/S ... ipting.ppt
Hello Jackus
Hello brasileo,
we had the same problem when we moved from CHS to EZCh. Agilent local representative made a macro for us which has functionality as S/N calculation in Chemstation. Everything is described in Powerpoint presentation (including necessary files).
Have a nice day.
http://rapidshare.com/files/453960098/S ... ipting.ppt
Can you please upload this file again? Thanks in advance.
Hello BrasileoThanks Jackus
http://www.ulozto.net/xGjPPRz/signal-to-noise-calculation-with-ezchrom-scripting-ppt
http://www.ulozto.net/xGjPPRz/signal-to-noise-calculation-with-ezchrom-scripting-ppt
Thanks a lot
http://www.ulozto.net/xGjPPRz/signal-to-noise-calculation-with-ezchrom-scripting-ppt
Thanks a lot
Hello again
I have 2 questions Brasileo,
Do yo have to make a special folder to put the scripts?
or
In Which Folder do you have to put them?
Thaks
Hello Nelsongar,
I cannot remember exact time it took. I suspect it was about up to two weeks. Finally the solution didn't come from Agilent but their local rep. who was programmer.
You can save files anywhere you want. You have to specify exact path as described in PPT. In our client/server version I saved them to server but you can save them locally into EZCh project folder (as example).
Thanks for your advice, but We need the calculations for S/N 6-sigma, because is part for validate methods in regulations for pharmaceutic industry, We are working in this problem, but the calculations are very incongruous and we are stuck in thatHi, you definitely can program scripts in OLCDS EZChrom, just like in the 'old' EZChrom. It is a bit different.
Furthermore, under Custom Parameters you have USP and EP S/N calculations. At least in versions A.04.04 A.04.04 and A.04.05. Please refer to the manual where it is exactly written how to perform such calculation. In one case you need a separate blank run to be acquired.