Minocycline
Posted: Sat Jun 30, 2012 12:09 am
by adam
Hey all
I was wondering if anyone can recommend a good method for analysis of Minocycline - separating it from its degs and from the epimer.
I did some quick searching and most of the methods I see use very atypical conditions, e.g. isopropanol in the mobile phase, unusual buffer systems, etc. I am not entirely sure what is the reason for this.
I think even things like HILIC or even classical normal phase may be a good alternative, since hydrolysis and epimerization occur much less readily in those types of mobile phases.
Well any push in the right direction would be appreciated.
Mark
Re: Minocycline
Posted: Sat Jun 30, 2012 4:55 am
by Vlad Orlovsky
might be a good candidate for our SHARC hydrogen-bonding columns:
http://www.sielc.com/Products_Sharc.html
http://www.sielc.com/Applications_By_Publish_Date.html
or mixed-mode.
If you send me samples I can look at this in the lab.
Re: Minocycline
Posted: Sat Jun 30, 2012 2:07 pm
by carls
Given all the functional groups present on this molecule I'm surprised relatively simple mobile phases can be used. I found a reference using the below conditions:
Mobile phase: 35 mM KH2PO4, pH7.5:TEA:ACN (74.5:0.5:25)
Column: Phenomenex Luna Phenyl-Hexyl analytical column (150x4.6 mm) with guard column
Detection Mode: UV detection at 280 nm
Flow rate: 1.0 mL/minute
Column temperature: 50 °C
Re: Minocycline
Posted: Sat Jun 30, 2012 6:31 pm
by unmgvar
c-18 with a ion pairing agent should work
Re: Minocycline
Posted: Mon Jul 02, 2012 6:01 pm
by adam
Thanks for the replies.
But what we were hoping for was a method already available that can separate Minocycline from its epimer. This is the hard part. It's not too difficult to get a method that can quantitate Minocycline, but getting one that can separate it from its epimer - that's the trick.
Re: Minocycline
Posted: Mon Jul 02, 2012 7:57 pm
by Vlad Orlovsky
Sending me a sample is always an option
Re: Minocycline
Posted: Mon Jul 02, 2012 8:30 pm
by Victor
Hi Vlad,
Would you classify the SHARC column as a type of normal phase column?
Re: Minocycline
Posted: Mon Jul 02, 2012 8:51 pm
by Vlad Orlovsky
it operates in normal phase conditions, but some additives are required to enhance or suppress hydrogen-bonding interaction.
Re: Minocycline
Posted: Tue Jul 03, 2012 2:17 pm
by carls
Here's a related reference:
http://www.basinc.com/library/presentat ... ffen02.pdf
The chromatographic conditions detailed along with example chromatograms gave separation of minocylcine from its epimer.