Simply cannot get the Thermo LTQ ion trap MS tune pass the manufacture's suggestion. No matter what we tries, the peah 524 (MARFA) is only at 30% of the base peak (supposed to be aound 60%). The tube lens voltage seems to be the major factore affecting the relative peak intensity for low (caffeine at 195) and high mass (Ultarmark at 1222, 1322, 1422, etc). If the tube lens voltage is near 100V, 195 is the base peak, near 200V, 1522 is the base peak, either way, peak 524 won't exceed 30% of the base peak.
New tube lens, new detector, new tuning solution. After installing a new electron multipler and restoring the factory calibration, re-calibrating th eystem failed to obtain acceptable mass and resolution calibration after quite a few tries. The service engineer seemed lost, so did I.
Any idea what could be wrong?