What is the reason behind the negative peak in chromatograms
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
2 posts
Page 1 of 1
In my method, mobile phase consists of 70% Buffer and 30% MeOH. Drug standards were prepared in Water (deionised). Then Why I am getting negative peak in first 3 minutes though the major portion of mobile phase if Water
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Without knowing what detector you are using (and if UV, what wavelength) and what buffer you are using, it could be any of several common things:
- With UV detection at short wavelength, deionized water may well have a lower absorbance than your buffer/methanol mixture.
- with UV detection at any wavelength, injection of a different diluent from your mobile phase will generate a refractive index anomaly (UV detectors do respond somewhat to RI).
- with any detector, the equilibrium upset from injection can make the baseline jump around ("t0 noise").
The high probability of t0 noise is one reason the US FDA suggests that k' values should be >2 for any peak that you quantitate.
- With UV detection at short wavelength, deionized water may well have a lower absorbance than your buffer/methanol mixture.
- with UV detection at any wavelength, injection of a different diluent from your mobile phase will generate a refractive index anomaly (UV detectors do respond somewhat to RI).
- with any detector, the equilibrium upset from injection can make the baseline jump around ("t0 noise").
The high probability of t0 noise is one reason the US FDA suggests that k' values should be >2 for any peak that you quantitate.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts
Page 1 of 1
Who is online
In total there are 19 users online :: 3 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot], Google [Bot], Semrush [Bot] and 16 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot], Google [Bot], Semrush [Bot] and 16 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science