UHPLC Resolution System Suitability Specification Revision?
Posted: Wed Jun 27, 2012 2:48 am
I have come across a few examples over the past few years where normally acceptable variability in retention times (<5%) gave resolution values that fail system suitability limits (Rs>2). It seems the small peak widths obtained with UHPLC (and, therefore, the smaller difference in retention times) places higher demands on batch to batch reproducibility and UHPLC instrumentation. Below is an example using gradient elution where a 2.5% (or 14s) change in relative retention for a compound dropped the resolution value from 2.6 to 1.6:

The smaller peak widths in case 1 (2% and 4%, respectively) do not compensate for the change in retention/selectivity in case 2 (~2.5%).
The fact that a relatively small change in relative retention (~2.5% or 14s) can reduce the resolution value by a large amount (~50%) makes me think system suitability criteria should be reviewed in light of UHPLC capabilities.
Any comments or opinions?

The smaller peak widths in case 1 (2% and 4%, respectively) do not compensate for the change in retention/selectivity in case 2 (~2.5%).
The fact that a relatively small change in relative retention (~2.5% or 14s) can reduce the resolution value by a large amount (~50%) makes me think system suitability criteria should be reviewed in light of UHPLC capabilities.
Any comments or opinions?