Hi
I agree installation is main problem. But split flow must be working also, and it needs to be at least > 1:5. Many of us do trust the GC readouts which can be wrong. Split lines can get blocked, same as split line filters. Check real the split flow using flow meters. If that is OK, we always use methane as a sample to see if installation is good. You need to see a sharp symmetrical peak. If this peak is OK, then you know installation should be fine.
Now if we inject solvent there still can be several issues causing tailing:
- is your injector hot? (check yourself.. also here don't "believe" your GC readout!
- is your solvent polar? If you use methanol, it will always give some tailing;
- is your column not activated? cutting a 20 cm section form inlet will tell you that.
- is your detector position OK? column should be close to the flame tip.
- also, if you use a coupling, many tailing problems happen within couplings due to improper column positining.
Don't forget to check for leaks using leak detector; once you have "sharp" peaks it does not mean there are no leaks. Leaks will cause activity, but usually only when analysis temperatures are >150C
rgsd
jaap de zeeuw, Restek corporation