unmgvar » Sun May 29, 2005 11:50 am
Hey Marck
your question actually touches many parameters. I hope that i will answer most of it's facet, especially the important ones.
1.the USP has created a special specification for columns. in order to create some orderin the chaos.
this set of specifications permits you to change column manufacturers if you stay within the system suitability criteria of your method.
2. both USP and EP have come out with a set of rules in order to be able to change columns if for exemple a certain type is no longer manufactured (this is for the common columns not the chiral). you can change column lenght, particle size and particle diameter if necessary, but only according to the rules that they have set.
3. EP has a chapter concerning what it regards to be correct peak shape and other parameters. USP has also a chapter like this but from what i know there are trying to implement a new set of rules that currently can be found in the forum and that serves has a guidelines for the industry.
for example according to EP, unless stated otherwise in the method, a correct peak shape should be between 0.8 to 1.5.
4. All and all ot os in the end up to you. you have to find, check, prove and validate your methods in order to prove robustness, repeatability, accuracy, precision and so forth. in the future you will only be able to accept mismatch that will be in accordance to your system suitability parameters. if you were flexible in the method validation then you will easily be able to switch column from different manufacturers.
if you have all the "goodies" in you system suitability RT, RRT, k', resolution, tailing, plate counts, S/N, and you can't achieve one of the list then you must stick to the column that will get you all those parameters.
hope this helps