We are running into issues between these 2 systems, reading through other posts it seems as though the only differences between the 1100 series and 1200 series are that the 1200 has a higher rating for pressure and compressibility. With the 1100 our chromatograms look just as they are supposed to. Our issue that we are seeing is with the 1200 the elution time is about 20% later than with the 1100 and much broater peaks. Same column was used, same flow cell, same flow rate, gradient, etc.
Sidenote: a 10uL injection on the 1200 gives us a perfect defined chromatogram, but at 20uL we are seeing a lot of fronting, suggesting sample overloading? Our method is validated for 20uL which worked fine with out 1100's but we are doing away with the 1100's. Any suggestions as to what the differences/issues we are seeing between these systems is greatly appreciated!