Clarithromycin raw material analysis

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Related substances of above material has to be determined using HPLC method as per USP 32. But this gradient doesn't work.
Can anyone help meto solve the problem
"doesn't work" covers a lot of ground. Can you be more specific about the method (we do not all have access to USP), what you have tried, and what the problem was?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
"doesn't work" covers a lot of ground. Can you be more specific about the method (we do not all have access to USP), what you have tried, and what the problem was?
Dear Sir,
Thank you for the reply and following is the test method.
Problem: no peak observed. Used with some adjustments, but we couldn't obtain a peak for high concentraions of standard solutions too.
Kindly advise me to solve the problem
Thank you

Clarithromycin - Related Substances
Chromatographic System:
Detector: 205nm, and a 4.6mm x 10cm column that contains packing L1 at 40 C. flow rate is ~ 1.1mL per minute.
Chromatograph program is as follows;

Time -- Solution A --------Solution B --- --------Elution
(Minutes) -----(%) --------- (%)

0 to 32 --- 75 to 40 ------------25 to 60 -------- Linear gradient
32 to 34 ------40 -----------------60 -------------Isocratic
34 to 36 ---40 to 75 ----------60 to 25 ------------- Linear gradient
36 to 42 ----- 75 ----------------25 -----------Isocratic

Relative retention time with reference to clarithromycin (retention time = ~ 11 minutes) include the following : impurity I = ~0.38; impurity C = ~0.89; impurity F = ~1.33; impurity A = ~ 0.42; impurity D = ~0.96; impurity P = ~ 1.35; impurity J = ~ 0.63; impurity N = ~1.15; impurity K = ~ 1.59; impurity L = ~0.74; impurity E = ~1.27, impurity G = ~1.72; impurity B = ~0.79; impurity O = ~ 1.38; impurity H = ~ 1.82; and impurity M = ~ 0.81.

Solution A:
4.76 g of monobasic potasium phosphate per L. Adjust pH to 4.4 with dilute phosphoric acid and pass it through C 18 filtration kit.
Solution B: acetonitril
Mobile Phase:
Use variable mixture of solution A and B as directed above.
Diluting Solution:
50:50 mixture of acetonitril and water
Standard solution A:
transfer 75mg of USP Clarithromycin RS to 50mL volumetric flask and dissolve in 25mL of acetonitril and dilute water to volume, and mix.
Standard solution B:
transfer 5.0 mL of std. solution A to 100 mL volumetric flask, dilute with Diluting solution to volume and mix.
Standard solution C:
transfer 1.0 mL of STD. solution B to a 10 mL volumetric flask, diute with diluting solution to volume and mix. it contains about 0.0075mg of USP clarithromycin RS per mL.
Standard solution D:
transfer about 15mg of USp clarithromycin identity RS, to 10mL volumetric flask and dossolve in 5.0 mL of acetonitril, dilute with water to volume and mix.
Test solution: Trasfer about 75mg of clarithromycin to a 50 mL volumetric flask and dissolve in 25 mL of acetonitril, dilute with water to volume and mix.
System suitability:
Chromatograph std. solution B and record the responses. tailing factor for the main clarithromycin peak is not more than 1.7.
Chromatograph std. solution D and record the responses. The peak to valley ratio of impurity D and clarithromycin is not less than 3.0
Procedure:
seperatly inject equal volumes of diluting solution, std. solution B, D & C and test solution.

Thank you
There's a troubleshooting guide on our web site that may be helpful:
http://www.lcresources.com/resources/TSWiz

In the meantime, here are some thoughts:

1. Did you see *anything*? Baseline drift, t0 noise, whatever. If not, then check to make sure you actually have flow in the system, that the detector is functioning properly (lamp is on, wavelength is correct, attenuation/range are correct), and that your injector is actually injecting something. At 205 nm you should be able to see a peak at or near t0 if you inject a 50/50 mix of water and methanol. If you still see nothing, re-run the PQ (Performance Qualification) on the system to confirm that the detector and autosampler/injector are functioning properly. While you're at it, check the gradient proportioning system to confirm that it is delivering the correct solvent ratios.

2. If you see a big peak at t0, it may be that all your peaks are either unretained or too strongly retained. If the peaks are coming out very early, your column may be inappropriate or may be dead. You gave the compendial procedure but did not say what column you actually used (all "L1" columns are *not* equivalent!). If you see only a small amount of t0 noise then, your peaks may be retained far past the end of your gradient. Try running a wider range gradient (from 5% to 90% organic) to check those possibilities.

3. If all else checks out, prepare fresh standards and mobile phases. I have been known to "slip a decimal point" and have my concentrations off by one or two orders of magnitude.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
4 posts Page 1 of 1

Who is online

In total there are 358 users online :: 1 registered, 0 hidden and 357 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google Adsense [Bot] and 357 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry