FID broken on Agilent 7890 GC

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
Whenever I put on my Agilent 7890A GC it always display "FRONT DETECTOR (FID) BROKEN". What could be responsible for this
Loose/disconnected cables.
Faulty electronic boards.
Which of the boards could this be so that I dont just grope in the dark
main mother board
FID electrometer board

best wishes,

Rod
Shortened heater.
Disconnected thermocouple.
Shortened to chassis collector electrode or dirt on collector.
5 posts Page 1 of 1