by
zheyin » Mon Jun 18, 2012 5:45 pm
The discharge tube from the excess solvent is well above the waste level. The exhaust tube from the back side is stretched over to a hood as expressed by the manufacturer.
I have heard that back pressure can cause issues. I have checked the flow test within the instrument and it passes. Is there another way to check the "vacuum" condition you refer too?
Also we have noted some poor performance with our nitrogen source, we have gone through two oxygen filters on a nearby GC. The ELSD uses the same nitrogen source. Can this impact the response issue?
The N2 flow rate is defined by your ELSD and has a guage to regulate, I belive its strength should stay constant during runs. There should be no significant particles in N2 to affect your result, either.
Do you have a meter to constantly monitor the vacuum strenth connecting to the ELSD?
Maybe the hood's sucking strength fluctuates...
When I used ELSD, I didn't have hood, but vacuum jack. The strength is not adjustable and too strong for ELSD, all I got was noise.
So I connected a buffering filtering flask in between, and it worked.