Empower ethernet detector not switching from ready>running

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
Empower 2154, SP/FR 4/D
2489 detector & 2414 hooked to a 2695. 2489 is ethernet, rest IEEE.
Last week, all was well. This week, using same method set, detector won't start collecting data on injection. It shows "Running" during monitoring and a baseline is displayed. When injections happen, it just stays on "Ready".

Any ideas?
Thanks,
DR
Image
Did your inject start cable become disconnected?
Nope - I checked that.

I also created a new version of the instrument using just the 2695 & 2489 (there is also a seldom used RI detector available on that system), wrote a new instrument method for it, included it in the method set... still no go.

I'm starting to think maybe it's a hardware fault on the detector.
Thanks,
DR
Image
3 posts Page 1 of 1

Who is online

In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry