Analysis of hyaluronic acid by reversed phase HPLC
Posted: Tue Jun 12, 2012 9:34 am
by archie
Hi all..
I am analysing hyaluronic acid in cosmetics.
I found many article that using GPC for that, but I can't afford this column.
So, is it possible to do it in usual RP column? I mean C-18 or C-8.. with 100-120A pore size.
The MW is about 1,400,000 Da.
I don't have wide pore column either.
Thanks in advance!
Re: Analysis of hyaluronic acid by reversed phase HPLC
Posted: Tue Jun 12, 2012 3:12 pm
by Vlad Orlovsky
we have a method for something similar (alginic acid), although it is smaller acidic polymer than you molecule:
http://www.sielc.com/Compound-Alginic-Acid.html
because of the polymeric nature of the molecule, in order to get a better peak you need to go into ion-exclusion/size exclusion mode. Please note that flow rate in our application is 4 times slower.
Posted: Wed Jun 13, 2012 10:38 am
by unmgvar
no you probably cannot do this compound on a standard RP column
you can maybe do an application with a wider pore column, like 300 or maybe better to try a 500 or 1000A pore volume.
even if you see something you'll probably have a very bad recovery on such a column
Posted: Wed Jun 13, 2012 3:10 pm
by Fernando
Archie
I´m sorry you have as all mortals go to the BP 2012 for example, and do it colorimetrically, especially if its a problem of money.
Good luck!
Posted: Wed Jun 13, 2012 5:07 pm
by Vlad Orlovsky
You can do it on smaller pore size columns in size exclusion mode. I think that even 300-500A might be too narrow for this polymer
Posted: Thu Jun 14, 2012 8:00 am
by archie
Ow..
I also find application that using ion pair reagent, but hyaluronic acid need to be derivatized first.
I don't have any experience in derivatizing sample either..
But I think it's rather complicated..am I right?
Posted: Thu Jun 14, 2012 6:06 pm
by zheyin
Ow..
Derivatization is like wet chemistry,
not complex at all...
just tedious!
Posted: Thu Jun 14, 2012 7:00 pm
by Vlad Orlovsky
What kind of derivatization they are talking about? It is hard to derivatize this polymer completely and you will end up with a horrible mixture of things. Go with size/charge exclusion at low flow rate and low UV, or use ELSD for detection.
Posted: Fri Feb 08, 2013 11:22 am
by max_planck
Is this working with a UV-detector at 210 nm for hyaluronic acid?
Posted: Fri Feb 08, 2013 1:25 pm
by Vlad Orlovsky
It is uv active Since it has a lot of carbonyls.
Posted: Tue Nov 26, 2019 7:48 am
by ydna1977
Hi ,
On the subject of Hyaluronic acid , what is the most suitable sample extraction technique to achieve suitable sample recovery 90-110% ie shake , vortex, stir , temp..? ..for various MW of HA for example 3-4 Mda in sample material/product
Posted: Tue Nov 26, 2019 12:37 pm
by ydna1977
Also what are the options for a force deg study on Hyaluronic acid with respect to ICH Q2 (R1) IE acid base oxidation light thermal etc and suitable conditions ?