differences in response
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
2 posts
Page 1 of 1
I have two fully qualified reference standards for a drug substance – first is a freebase and second is a TFA salt. The test method is a regular HPLC on C18 column with 0.1% TFA/Water/Acetonitrile mobile phase, gradient program up to 75% organic, and detection at 215 nm. The response of freebase standard (area/concentration) is about 104% of the response of TFA salt standard (converted on freebase for calculations of concentration). I have no reason to suspect an overestimation of purity value of the TFA standard because it was investigated and corrected for all organic and inorganic impurities, moisture content and residual solvents. I may suspect that there is an impurity (with higher UV absorption) hiding under the parent peak in freebase standard although it is highly unlikely. What else could cause this phenomenon?
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
Any differences in peak shape and/or retention time can exacerbate the random variation expected between different preparations of the same standard material.
Using different salts will often result in larger differences in peak shape and retention time than what you get from random variation, so I would think that 104% is not bad (since, to a small extent, you are comparing apples to oranges).
If you were to work the free base up in a solution of TFA so that your final dilution of it chemically matched the composition of the tfa salt preparation, I'd expect a smaller residual difference between the 2.
Using different salts will often result in larger differences in peak shape and retention time than what you get from random variation, so I would think that 104% is not bad (since, to a small extent, you are comparing apples to oranges).
If you were to work the free base up in a solution of TFA so that your final dilution of it chemically matched the composition of the tfa salt preparation, I'd expect a smaller residual difference between the 2.
Thanks,
DR

DR

2 posts
Page 1 of 1
Who is online
In total there are 9 users online :: 0 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 9 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 9 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science