This did not solve the pressure problems however. Replaced pump seals. Didn't solve the problem either.
Interestingly, each pump head independently seems to pump just fine and can sustain accurate flow rates, even through the system (but bypassing column). However, when both pump heads are connected to the outlet manifold, only one will pump (i.e. solvent flows out only when one piston is moving out).
I've checked and rechecked for loose/leaking inlet and outlet connections, and even taken apart the pressure transducer/outlet manifold. Solvents are degassed HPLC grade.
At this point I'm out of ideas. Any thoughts?
Thanks for any help.