No you didn't miss anything, in fact as I re-read the whole dialog I see that *I* missed what you were driving at at first.
Still, I can see two limiting cases:
- UHPLC is used to improve speed at a given resolution (and pressure). In that case, the plate counts will about the same as in an HPLC method (somewhere around 10k for a good column on a good day), but they will be generated in less time by a shorter column, and the selectivity constraints will remain about the same. This, by the way, has characterized the evolution of HPLC over the past four decades: from 30-cm long columns with 10-micron packings generating about 10k plates in 20-40 minutes to 10-cm long columns with 3 micron packings generating about 10k plates in 3-5 minutes (in both cases for simple isocratic separations).
- UHPLC is used to determine resolution via increasing the plate count. In that case the run times will stay about the same, but somewhat longer columns will be used. Getting *huge* plate counts is still impractical on a day-to-day basis, with a doubling (to about 20k plates) about what you can expect (yes, you can get into the 40k range, but that's at the bleeding edge). A back-of- the envelope calculation suggests that for Rs = 2.0, a 10k plate column will need an alpha value of 1.08, and a 20k column will need an alpha value of 1.06.
Here's the table of required alpha values for a given resolution and plate count:
At this point, I suspect that 20k plates is about as high as most people will go, so the vulnerability to selectivity variations is there, but isn't catastrophic. You're right that it's a tighter constraint, but that's just one more of the "gotchas" with the small-particle columns (to go with fittings and filtration!).