AICMM » Wed May 09, 2012 2:10 pm
cfnichol,
First, you are really going to want to install valves to do the injections if you want to do this right. You're ready for that, right? Second, an HID (PDD by Valco or DBD by my company, AIC) can replace the FID to get you low concentrations of CO, CO2, and methane. I have not seen a Gow Mac HID as a stand-alone product for other GC's. Or you can install the methanizer before the FID to get not as low levels of CO, CO2. Cost will probably be pretty comparable actually. TCD will probably not get you low enough.
The other thing you need to take into account (in switching from capillary to packed) is that your current flow control is for splitless/split and you are moving to packed. So, at a minimum, you will need to install packed pressure controllers.
If you want to drop me a line, I can talk your ear off. aicmm at flash.net
Best regards,
AICMM