Thanks, Carras and Jody, following your advices I've managed to recalibrate the instrument successfully.
But having solved one problem, I've got a couple of new
1. I've made the compound optimization (with Autotune - developing MRM), everything went OK. But then Autune started giving strange results - it sets Cone Voltage for any compound to 100 V (maximum in the ramp). If I limit ramping to any other value - Autotune gives this value like optimum. Although when I ramp Cone Voltage manually, hte highest intensity is reached at cone voltage far from 100 V (30-45 usually for different compounds). I wonder why Masslynx offers me maximum anytime.
2. The second is even worse. At the end of day the Source Temp readback SUDDENLY began to increase (I set 120 C). I found it on 160 C, then it increased and increased, and I thought I better vent and turn off. By this time the readback showed 520 C (!!!). Is it possible in principle? I checked the probe and source cover, they seemed not very hot (one could touch by hand). So I turned the instrument off and went home completely embarrassed.