problem with calibration review in EzChrom Elite

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

4 posts Page 1 of 1
Thanks in advance.

Having trouble getting the software popluate peak area information into the calibration review window. The intergration is good, peak is recognized in the chromatogram. The calibration review winsow takes in the concentration/levels specifided in the peak/group section, but not the peak area information. Any one know what's the problem?

Agilent is very lazy in responding to service request. My next GC, I believe, won't be an Agilent one.
Did you run the sample as a calibration sample from the sequence. Did you specify the level then in the sequence table. Or did you just run it as a single sample?
I need a bit more information from you to find out what step is missing in the calibration.
Freek Varossieau
OpenLab CDS 2 specialist
BeyondOpenLab
beyondopenlab@gmail.com
+5977114721
Thanks in advance.

Having trouble getting the software popluate peak area information into the calibration review window. The intergration is good, peak is recognized in the chromatogram. The calibration review winsow takes in the concentration/levels specifided in the peak/group section, but not the peak area information. Any one know what's the problem?

Agilent is very lazy in responding to service request. My next GC, I believe, won't be an Agilent one.
Please contact with me, I'll help you anything about EZchrom. I can control by teamviewer to solve for you.

skype: tuanhappy
Freek is right as he should be coming from the source itself. You will see your peak area info in the table under the Review Calibration Table once you have entered your Peak info in the Peaks/Groups table. Many people forget this as its needed for the calibration table to populate. Then use the Run Type Calibration Begin and add your samples level by level. Be sure to also have your sequence table pointing to the right method and data folders. This can be found in the Sequence menu under the Properties window and then File Paths. Once this has been done, process the sequence and the table will populate. Remember that each calibration is method file specific.

BTW, Having an Agilient rep on this board shows that they are not lazy at all. They keep track of most issues in the chromatography arena
4 posts Page 1 of 1

Who is online

In total there are 131 users online :: 1 registered, 0 hidden and 130 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google Adsense [Bot] and 130 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry