purge valve won't close?
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
9 posts
Page 1 of 1
I have an Agilent 1200 HPLC with quad pump. After opening the purge valve to prime, I can no longer re-direct the solvent through the system by closing the valve. Instead 90%+ of solvent continues to bypass the system and go into waste line even when purge valve appears closed tightly. I dont note any stripping of the threads on the valve knob, and it turns with a level of ease similar to before the problem. Any suggestions for fixes? I've considered replacing the valve frit, but wonder if thats really a possible culprit, since the manuals indicate a bad frit leads to increased pressure -- which I haven't seen. Thanks for advices!
-
- Posts: 106
- Joined: Wed Nov 23, 2011 11:35 am
replace the valve completely..
-
- Posts: 762
- Joined: Thu Dec 13, 2007 5:54 am
I agree, sometimes purge valves fail. Agilent sells the complete assembly. You might want to disassemble the valve as much as possible and see if there isn't something in there preventing a complete seal, but I would only do that when I have a new purge valve on the instrument. Never hurts to have a backup if you find some sort of obstruction in the valve (which is doubtful, but worth a shot once you've replaced it).
Time flies like an arrow. Fruit flies like a banana.
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
New purge valve assembly. Oder two, keep one as a spare, we do.
-
- Posts: 377
- Joined: Wed Aug 24, 2011 2:10 pm
Are they as rubbish as the ones on the 1100?
Where can I buy the kit they use in CSI?
-
- Posts: 100
- Joined: Thu May 21, 2009 7:50 am
Mine works fine. Four years of using 1200.
-
- Posts: 345
- Joined: Sun Dec 06, 2009 6:16 pm
Have you replaced the gold seal (PN: 5001-3707)? The "frit" is a cylidrical filter and is not part of the sealing mechanism
A. Carl Sanchez
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Same.Are they as rubbish as the ones on the 1100?
Rule #1: don't overtighten
Rule #2: don't overtighten
Rule #3: guess !!!
-
- Posts: 1
- Joined: Wed May 30, 2012 6:49 am
Hey I also had the same problem. I use Agilent 1260 Quaternary pump. I dismantled the purge valve from the knob side by just putting the lock out. I unscrewed the whole knob out and found a lot of particles inside, I care fully cleaned it with HOT water and IPA. Assembled back. Now its working.
But do it only if u have experience over it. Make sure u dont loose any minute parts while opning the knob.
But do it only if u have experience over it. Make sure u dont loose any minute parts while opning the knob.
9 posts
Page 1 of 1
Who is online
In total there are 26 users online :: 1 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 25 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 25 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science