purge valve won't close?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

9 posts Page 1 of 1
I have an Agilent 1200 HPLC with quad pump. After opening the purge valve to prime, I can no longer re-direct the solvent through the system by closing the valve. Instead 90%+ of solvent continues to bypass the system and go into waste line even when purge valve appears closed tightly. I dont note any stripping of the threads on the valve knob, and it turns with a level of ease similar to before the problem. Any suggestions for fixes? I've considered replacing the valve frit, but wonder if thats really a possible culprit, since the manuals indicate a bad frit leads to increased pressure -- which I haven't seen. Thanks for advices!
replace the valve completely..
I agree, sometimes purge valves fail. Agilent sells the complete assembly. You might want to disassemble the valve as much as possible and see if there isn't something in there preventing a complete seal, but I would only do that when I have a new purge valve on the instrument. Never hurts to have a backup if you find some sort of obstruction in the valve (which is doubtful, but worth a shot once you've replaced it).
Time flies like an arrow. Fruit flies like a banana.
New purge valve assembly. Oder two, keep one as a spare, we do.
Are they as rubbish as the ones on the 1100?
Where can I buy the kit they use in CSI?
Mine works fine. Four years of using 1200.
Have you replaced the gold seal (PN: 5001-3707)? The "frit" is a cylidrical filter and is not part of the sealing mechanism
A. Carl Sanchez
Are they as rubbish as the ones on the 1100?
Same.

Rule #1: don't overtighten

Rule #2: don't overtighten

Rule #3: guess !!!
Hey I also had the same problem. I use Agilent 1260 Quaternary pump. I dismantled the purge valve from the knob side by just putting the lock out. I unscrewed the whole knob out and found a lot of particles inside, I care fully cleaned it with HOT water and IPA. Assembled back. Now its working.
But do it only if u have experience over it. Make sure u dont loose any minute parts while opning the knob.
9 posts Page 1 of 1

Who is online

In total there are 26 users online :: 1 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry