Reduced velocity - effect of temperature

Basic questions from students; resources for projects and reports.

3 posts Page 1 of 1
Hi folks,

I have a question relating to the reduced velocity (v) of the mobile phase, as defined by the equation:

v = u * dp/Dm,

where u is the average interstitial velocity of the eluent, dp is the particle diameter of the packed column and Dm is the diffusion coefficient for the solute.
When increasing temperature, we increase the diffusion coefficient of the solute within the mobile phase as per the Wilke-Chang equation. The corresponding effect is to reduce the magnitude of the v term as per the above equation. My question, therefore, relates to how increasing the flow rate can compensate for this effect and so return us to an optimum value for h (the reduced plate height), offering maximum chromatographic efficiency? I’m interested to understand, at a mechanistic level, what exactly is causing the value of v to decline when temperature is increased?

From what I can deduce, there are two possible options:

1) A reduced velocity of solute diffusion in to the particle pores – though with an enhanced diffusion coefficient, I’ve no idea how that could make sense as I’d imagine molecules to move more freely in to the mobile phase within the pores.
2) A reduction in the B term of the Knox equation resulting from use of a greater flow rate.

Option 2 makes the most sense to me personally, however, due to its absence in the above equation, I wonder whether I have perhaps misunderstood.

Option 1, meanwhile, confuses me somewhat. If we were to have a reduced velocity of solute diffusion in to the pores of a packed column’s particles, would we not expect an increase in flow rate to enhance the spreading of a solute band? My thoughts on this come from the fact that we are considering an equilibrium processes whereby, as mobile phase passes the stationary phase, some of the solute molecules are retained within/on the stationary phase (resistance to mass transfer of the stationary phase). These retained molecules lag then behind what will eventually become the apex (point of highest concentration) of a chromatographic peak within the detector. By increasing flow, I imagine they would be left even further behind – so creating even greater spread and hence, reduced efficiency.

As a final part to the post, could someone perhaps confirm or deny whether or not by increasing flow rate in response to elevated temperature, we might improve efficiency, but in so doing, reduce our resolution? I’d imagine it’d be a trade off between two events: 1) narrower peaks that provide an improved resolution if the retention values remained unchanged; 2) the higher flow rates would lead to a reduction in retention factor (k), in turns means reduced separation of analytes.

All the best,
U0mrj1K
In practice almost no one works at the optimum velocity or even tries to do so.

First of all, given that resolution depends on the square root of the plate number, method development time/effort are much better invested in manipulating selectivity.

Second, the loss of resolution resulting from decreased k' can always be compensated by adjusting the mobile phase strength to keep k' constant.

Why does the optimum velocity change with diffusivity? Take a look at the van Deemter equation (or the Knox equation if you're doing LC and are a stickler for detail). Increasing diffusivity increases the B term (molecular diffusion) and decreases the C term (mass transfer). That shifts the minimum (optimum) to the right (higher velocity).

If you *really* want to get into the theory of chromatography, get a copy of Giddings' classic "Dynamics of Chromatography" (http://tinyurl.com/7l4b85u ) and wade through the math.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thank you Tom and apologies for harping on about the optimum values once again!

I'm a rather inquisitive chap and am looking to ensure my theoretical understanding is (as) solid (as possible) before undertaking some 'real-world' experiments.

Your recommendation for Giddings' book is much appreciated. It looks a brilliant text!

All the best
3 posts Page 1 of 1

Who is online

In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry