Is there a split vent valve on a GC?
Basic questions from students; resources for projects and reports.
12 posts
Page 1 of 1
Does anyone know anything about how the split and split vent actually works? Is there a valve in there? If so, do you have a schematic of it someplace I could look at? My boss suggested that the split wasn't working well because I wasn't getting reproducible results and I cleaned and replaced the tubing today –it was quite clean surprisingly. The split vent filters are backordered about two weeks, so I’m waiting on that. I’m wondering since the tubing was so clean with methanol rinses that perhaps the split material is just not reliably going out. I replaced the tubing anyway just in case. Any thoughts would be helpful. thanks.
-
- Posts: 64
- Joined: Mon Dec 05, 2011 11:41 pm
Can you specify what type of GC you have?
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
All of the major GC manufacturers have diagrams of their split systems on their websites, some of them have animations of where the gas goes and what valves are open and closed.
Peter
Peter
Peter Apps
-
- Posts: 25
- Joined: Mon Jan 09, 2012 9:31 pm
My GC is an Agilent 6890.
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
If you have a split/splitless inlet, there is a split valve. The PTV inlet has a similar valve - and off the top of my head, I don't recall if it gets a different name in this inlet.
The information available on the front panel should tell you what kind of inlet you have. And, if you can set a split ratio, you have a split valve.
The information available on the front panel should tell you what kind of inlet you have. And, if you can set a split ratio, you have a split valve.
-
- Posts: 25
- Joined: Mon Jan 09, 2012 9:31 pm
okay.
thank you.
I looked into this and found that the inlet liners i have are for splitless injection because this is what Agilent sells as UltraInert and I was using them and doing both splitless and split injection work with it depending on the sample. Apparently the splitless liner doesn't work well or repeatedly if you are doing split injection method.
I had wanted to use single taper inlet liners for split injection but Agilent's engineers tell me there is no such thing as using single taper inlet liners for split injections.. is this true? I had thought that previously I had seen this used.
So I think the answer is that I have to change the liner if I change from split to splitless method. Is this true?
thank you.
I looked into this and found that the inlet liners i have are for splitless injection because this is what Agilent sells as UltraInert and I was using them and doing both splitless and split injection work with it depending on the sample. Apparently the splitless liner doesn't work well or repeatedly if you are doing split injection method.
I had wanted to use single taper inlet liners for split injection but Agilent's engineers tell me there is no such thing as using single taper inlet liners for split injections.. is this true? I had thought that previously I had seen this used.
So I think the answer is that I have to change the liner if I change from split to splitless method. Is this true?
-
- Posts: 287
- Joined: Fri Jul 16, 2010 5:47 pm
Yes it is true. You can use a tapered liner in split mode for single analysis diagnostic type run but it won't be very reproducible as the split flow has to flow out the channels of the gold seal at the bottom of the inlet. The taper doesn't allow for this to flow very well. At least this is my experience. I just wouldn't do it. Split liners are usually pretty cheap. If your compounds aren't highly reactive then you probably don't need the most expensive type of deactivated liners either. This can really run up the cost. Simple 4 mm ID split liner should work fine.
~Ty~
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Having done split and splitless injections on ordinary taper-at-the-bottom Agilent liners my experience has been that you can do split injections to a splitless liner, but sometimes not vice versa since the narrower split liner cannot accommodate all the vapour from a splitless injection.okay.
thank you.
I looked into this and found that the inlet liners i have are for splitless injection because this is what Agilent sells as UltraInert and I was using them and doing both splitless and split injection work with it depending on the sample. Apparently the splitless liner doesn't work well or repeatedly if you are doing split injection method.
I had wanted to use single taper inlet liners for split injection but Agilent's engineers tell me there is no such thing as using single taper inlet liners for split injections.. is this true? I had thought that previously I had seen this used.
So I think the answer is that I have to change the liner if I change from split to splitless method. Is this true?
The critical test of whether you need to be buying specific liners is whether you can get the performance you need from the liners that you have - if you can then save your money. The presence or absence of glass wool in the liner will have a much greater impact.
Peter
Peter Apps
-
- Posts: 377
- Joined: Wed Aug 24, 2011 2:10 pm
FWIW the split vent filters are just some charcoal I think, so that you don't get gassed by the traces of whatever you're analysing.
Where can I buy the kit they use in CSI?
-
- Posts: 25
- Joined: Mon Jan 09, 2012 9:31 pm
I replaced the split vent filters. That wasn't the problem though.
I still can't get reproducibility within runs. What troubles me the most is that in a mixture of flavor components I get varying RSDs for each component. t he percentages are anywhere from 2 to 48% for an individual component within the overall flavor say for example something like orange. If I run the same orange sample liquid injection seven times, and then look at each peak, the separate components that I know make up the flavor since we made it in house don't have the same peak areas. Any thoughts as to why this might be?
I still can't get reproducibility within runs. What troubles me the most is that in a mixture of flavor components I get varying RSDs for each component. t he percentages are anywhere from 2 to 48% for an individual component within the overall flavor say for example something like orange. If I run the same orange sample liquid injection seven times, and then look at each peak, the separate components that I know make up the flavor since we made it in house don't have the same peak areas. Any thoughts as to why this might be?
-
- Posts: 287
- Joined: Fri Jul 16, 2010 5:47 pm
What is the history of your current GC method? Has it worked in the past or are you starting from scratch? Did it just start "acting up"? Let us know some more details (i.e., injection volume, injection temp, solvent oven start temp, injector pressure, flow rate etc.)
~Ty~
-
- Posts: 1
- Joined: Sun Jun 17, 2012 4:06 am
Yes it is true. You can use a tapered liner in split mode for single analysis diagnostic type run but it won't be very reproducible as the split flow has to flow out the channels of the gold seal at the bottom of the inlet. The taper doesn't allow for this to flow very well. At least this is my experience. I just wouldn't do it. Split liners are usually pretty cheap. If your compounds aren't highly reactive then you probably don't need the most expensive type of deactivated liners either. This can really run up the cost. Simple 4 mm ID split liner should work fine.
thank you for the information! i really appreciate it a lot!
12 posts
Page 1 of 1
Who is online
In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 18 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 18 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science