volatile fatty acid analysis
Posted: Tue May 29, 2012 1:00 am
by sepscientologist
I'm given aqueous samples of VFA (volatile fatty acids) which are pH 6 ish. My standard used to be just about 5 mg/mL of the acids
in water and hence, about pH 2. When I adjust the pH of my standard to pH 6 the response factor goes way down! Duh I guess it
would.
Should I acidify all my real samples and if so, what acid and what pH? I'm using a PEG type column.
Or, can I esterify? I have tried a few alcohols but the alcohol always seems to coelute somewhere around the Propionic acid ester.
Re: volatile fatty acid analysis
Posted: Tue May 29, 2012 2:02 am
by chromatographer1
I have seen in the literature, extractions of free acids into ether using nitric acid.
I have also seen direct aqueous injection using sulfuric acid.
I don't remember the details. Perhaps a trip to the library or a search on the internet might be of great benefit.
best wishes,
ROd
Posted: Tue May 29, 2012 8:09 am
by Peter Apps
VFAs in water are a hardy perrennial on the forum - have a look through the archives. What are you injecting ?, an extract or the aqueous matix ?, or headspace ?
Posted: Wed May 30, 2012 3:41 pm
by sepscientologist
injecting aqueous samples containing short chain fatty acids. Results are useless because the pH
of the samples is sometimes below and sometimes above pKA'a.