HPLC method for sara analysis

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hi everyone,
I am using Infinity 1260 series by agilent. I do not have HPLC method for sara (saturates, aromatics, resins and asphaltene) analysis of crude oil. I need a HPLC method for sara anylsis, can you people help me out?

thanks and regards
Akbar
I suppose your sample is crude oil. First dissolve the sample in hexane to precipitate out asphaltenes from the rest and recovered it by filtration.

Inject the saturates and aromatics onto NH2 column with hexane as mobile phase. Saturates and aromatics peaks can be quantified.

Resin is retained strongly and did not elute with hexane. 30%DCM/hexane backflush needs to be used to elute resin.

You can refer to the article
"Rapid and Accurate SARA Analysis of Medium Gravity Crude Oils" for details.

Hope this helps.
Thanks vnaren,
Yes I want to do same, but I did not run the analysis on this intrument as I do not have parameters to set it. I also need to know how to set back flush and which column we need to install for this?
Can you guide me in this

Kind regards
3 posts Page 1 of 1

Who is online

In total there are 17 users online :: 3 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Google [Bot], Semrush [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry