- Posts: 3
- Joined: Tue May 01, 2012 7:57 pm
I am currently in the process of developing a method to quantify ergosterol (converting to fungal biomass) in leaf litter, performed with an Agilent 7890A gas chromatograph interfaced to an Agilent 5975C inert mass-selective detector, containing an HP-5MS (30m x 0.250 mm I.D., 0.25 μm film thickness) capillary column. Standards and samples were injected in splitless mode at a volume of 1 μL with a starting oven temperature of 100° C. However, I have a set sub-samples that were sent to another lab in Mississippi and analyzed using HPLC and the comparative results indicate that the sensitivity of the GC-MS drops after reaching over approx 100 ng/uL. My faculty sponsors are ecologists and are not analytical chemists and they have suggested that I re-run my samples in split mode (1:10). I do not see the benefit of this. I feel like my samples would have even lower abundances than they already do (~500), therefore becoming un-quantifiable. I have few resources to talk with about this and I was hoping someone out there could help! Would my ergosterol concentrations be lower than they already are if I turn on split mode? please help!! Thank you so much!!!
-Caitlin Reece