Clarus 500 FID jet assembly
Posted: Wed May 23, 2012 9:52 am
by azharrashid
Dear all,
I have Clarus 500 GC with FID. Few days back I had to take out the FID jet assembly for cleaning. While doing so the ceramic tip was broken. I am in serious trouble for this. i shall be thankful if someone could give tips for how could repair the broken ceramic tip.
Azhar
Re: Clarus 500 FID jet assembly
Posted: Thu May 24, 2012 5:37 am
by dblux_
It's a common practice to replace broken FID jets.
Posted: Mon May 28, 2012 6:15 am
by azharrashid
Thanks for reply. Actually, I am out of budget at the moment and looking to repair the proken tip. Let me know if you have any idea about that.
Thanks.
Posted: Mon May 28, 2012 3:00 pm
by Don_Hilton
Check on the cost of a new jet. These are consumables, like inlet liners and columns -- even though they are not replaced as frequently.
On repair: These have to endure high tempeatures, must be aligned properly, and can not bleed solvents or monomer because this would affect the signal in the flame.
Posted: Mon May 28, 2012 4:00 pm
by dblux_
Thanks.
"Poor's man" solution is to use "water glass" (
http://en.wikipedia.org/wiki/Sodium_silicate ) as adhesive.
PS - use only very thinny amount, otherwise you will clog orifice