Hi. I'm new to this forum and i hope I'll find an answer to my question. I have SFC unit that i use to quantify olefins in gasoline. Lately I see a lot of peak tailing just for olefins. Tried changing temperature, flow, pressure .. nothing helps. Strange thing: it only does that on synthetic samples ( i.e. mix of 3-4 different hydrocarbons) and not on regular gasoline. Please help.
Re: SFC Olefins problems
Posted: Sat May 19, 2012 6:47 pm
by carls
the peak shape for alkanes is OK?
What columns are you using?
Re: SFC Olefins problems
Posted: Tue May 22, 2012 5:15 pm
by madchemist83
Alkanes and aromatics are fine. Also olefins peak only trailing in performance mix but not in actual gasoline sample. Silver column is from Selerity
Re: SFC Olefins problems
Posted: Wed May 23, 2012 2:18 pm
by carls
Have you prepared and/or purchased a fresh standard? What is the water content of the major components in your standard? This behavior points toward a difference in the matrices/solvent strength.
Re: SFC Olefins problems
Posted: Wed May 23, 2012 9:19 pm
by madchemist83
Performance mix is in-house made about a year ago. I'll try to make a new one. All components are hplc grade and shouldn't have much water in them
Re: SFC Olefins problems
Posted: Wed May 23, 2012 10:10 pm
by madchemist83
made another mix. same results.
Re: SFC Olefins problems
Posted: Thu May 24, 2012 3:40 pm
by carls
I may have barking up the wrong tree here. Its been many years since I ran this analysis and I was (mistakenly) thinking the silver column was before the silica column but it is the opposite.
However, even non-polar polar solvent like heptane can have as much as 100ppm water in them.
Is it possible the additive(s) in the gasoline samples are improving peak shapes? Perhaps its time to try another silver column.
Is the olefin mass load similar for both the standard and the samples (trailing peak(s) means tailing or fronting)?
Re: SFC Olefins problems
Posted: Thu May 24, 2012 8:23 pm
by madchemist83
Those chemicals do have water in levels 0.005% so that's 50 pppm. Will that affect chromatogram ? Standards olefin peak under same conditions will usually have about 0.5% less olefins then olefin peak of actual samples.
Re: SFC Olefins problems
Posted: Thu May 24, 2012 11:14 pm
by carls
Water is a strongly adsorbed on silica in normsl phase/ SFC so yes the water can affect the chromatography . I would expect the retention times of the alkanes/aromatics to be more affected. However the water could make it to the silver column and affect retention/chromatography there as well
Re: SFC Olefins problems
Posted: Fri May 25, 2012 12:33 am
by madchemist83
Can you recommend any way of cleaning silver column?
Re: SFC Olefins problems
Posted: Fri May 25, 2012 12:46 am
by madchemist83
Also how would I flush column with cleaning chemicals ? The only way I think is to inject it like a sample but it only injects small amounts
Re: SFC Olefins problems
Posted: Fri May 25, 2012 1:51 pm
by carls
I would contact selerity for their recommendation for flushing/cleaning-up the silver column. The conditions differ depending on the the type of silver loaded column used (silica or silica + SCX).