- Posts: 6
- Joined: Tue Mar 20, 2012 1:09 pm
I see this with two method with two different columns. I am assuming it is the same stuff (pluronic f68). One method: 90%water10%thf to 100% thf on DVB column. If I run 50%water50%thf to 100% thf I see no peak. I believe this is because the mobile phase is strong enough to not allow it to build on the head of the column. My other method is 100%meoh to 100%thf, also on DVB column. I tried a brand new DVB column and still see stuff! (different pore size, so it looks a little different) Using ELSD detection, peak is at about 90%thf, 10% meoh.
I do believe the contamination is before the column since my retention times are always the same.
I do not believe it is in the injector because: I took out the needle and needle seat and sonicated them in meoh. While I did this, I hooked the pump directly up to the column, so no part of the injector was in the system. I continued to do runs and still got my peak.
I ran a special cleaning mix of organic solvents in all mobile channels up to where the column is (kept the column off) for 20 min, 5 mL/min. Cleaned my mobile bottles (with the fresh hplc grade solvent straight from a virgin bottle) and refilled them. Then ran my solvents through the lines. changed the purge valve frit. Still seeing the dang thing! I don't know where else it could be!
Any thoughts?