- Posts: 9
- Joined: Mon May 21, 2012 9:36 am
I have a question relating to a sentence I have recently read in Synder, Kirkland and Dolan's (2010) book entitled: Introduction to Modern Liquid Chromatography, 3rd ed., page 42.
The sentence followed the equation: W^2 = ([16/L]*tr^2)*H; where W^2 represents the sum of the intra-column contributions to band broadening (Eddy diffusion/longitudinal diffusion/resistance to mass transfer in both mobile and stationary phases), tr is retention time, L is column length and H is the height equivalent to a theoretical plate.
I quote:
Am I right in thinking that values for H can be regarded as approximately independent of tr, because the contribution by the longitudinal diffusion is (relatively) small under "normal" HPLC conditions? I think I'm confused as to why the authors have decided to state that tr has seemingly no effect on H, when my understanding was, that it does.where values of H = L/N for different solutes are approximately independent of retention time tr for a given column of length L and the same experimental conditions.
All the best,
U0mrj1K