0.08 ug/l detetcion limit requirement for VOCs

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

5 posts Page 1 of 1
We are being asked to have PQLs of 0.08 ug/l for most VOCs by EPA 624. Any suggestions?

Thank you.
Just say no.
Realistically, while some targets may achieve an MDL of 0.08 ug/l on a clean system, unless your samples are very clean and something close to magic happens you won't see a PQL that low especially across the board. This is a wastewater method after all. Drinking water maximum contaminant levels are higher than that, at least in the US.
Just say no.
Realistically, while some targets may achieve an MDL of 0.08 ug/l on a clean system, unless your samples are very clean and something close to magic happens you won't see a PQL that low especially across the board. This is a wastewater method after all. Drinking water maximum contaminant levels are higher than that, at least in the US.
I'm working with VOCs on water recently and i couldn't say it better. He's right, even if you can go that low on concentration there is too much contamination risk to obtain good performances, at that level.
Davide Balbo from Italy
Who's asking and to what end. I'm running 524 and could not get down to 0.08!
The industry would need to move to a more expensive approach to achieve these limits. The current approach is limited where we strip the VOAs out of 5 to 25 ml of water, then discard 90 to 99 percent of the analyte in order to introduce them to the detector. Things have really gone backwards in this respect.

The approach used for air cans retains all the sample, but uses relatively expensive and balky techniques to remove the water and carbon dioxide. Of course, at lower levels, absorbtion and carry-over are bigger problems than detector sensitivity. There is a lot of crappy data out there for some of the heavier analytes.

Better techniques exist, but they are more expensive and cannot compete with the lab that justifies reporting the latest lower limits with an unrealistic MDL study. It all gets a bit frustrating.
5 posts Page 1 of 1

Who is online

In total there are 42 users online :: 0 registered, 0 hidden and 42 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 42 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry