Page 1 of 1

Electronic parameters of Varian star 3400cx

Posted: Thu May 17, 2012 7:26 pm
by mechemunioz
Hello,
I need some technical details of a Varian star 3400 cx GC.
I want to know the variation range of the output signal of the chromatograph (mV) and what is the interval of time between two points of acquisition.
Thanks.
Mercedes

Re: Electronic parameters of Varian star 3400cx

Posted: Fri May 18, 2012 12:28 am
by AICMM
mechemunioz,

Typically the 3400's had a 0-1 mV output for strip charts, a 0-1V for some data systems and a 0-10V output for the data systems that could use that. Signal out was analog so there is no data rate, that would be set by whatever analog to digital converter you were using.

Best regards,

AICMM