- Posts: 15
- Joined: Wed Oct 05, 2011 6:44 pm
I have an old system that I've brought back from the dead, but am now starting to have issues. I was getting air bubbles in the line after the Gradient Proportioning valve and before the pump. I checked my helium sparge and everything seems to be working fine there. My GPV has never worked quite right with the "A" reservoir never pumping, but "B", "C", and "D" worked fine.
I opened up the pump housing to see if there were any leaks around the GPV and saw none. Checked all the connections, and, now, when I go to prime the pump, only the "B" reservoir will prime! I can't even get liquid to the the pump on the "C" and "D" reservoirs.
I switched the GPV out for a spare one that came with this system from another division in our company and the "B" won't prime on that one, but "C' and "D" are working fine. I went online to look at replacement GPVs only to find them listed at $3000.
My questions are: Can I send the GPVs I have out somewhere for rebuilding (I can't find it anywhere on Waters' website)? Once they are rebuilt, will that help with the air in line issue?