Ultimate3000 Flow manager error

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hy everyone!
I am having difficulty with the system Dionex Ultimate3000. On the panel of Flow manager appears error "Flow control valve failure" or " Flow control valve limit" on panel. In Cromeleon Log file error is "flow control valve failure or column path not tight" or 'insufficient pump flow". When I disconnect column from switching valve, there is no error, when I reconnect column appears error message. When error appears the running program is aborted. Chromatograph operated by HyStar. I washed with isopropanol system to avoid the presence of bubbles, pressure in the system does not jumps. What could be the cause of the error?

Alena
Hi Alena,

Have you tried running the Diagnostic test within Chromeleon? If you are running version 6.8 SR2 or later, this should be available in the control panel menu under Control>Diagnostics. This may provide you with a little more information as to the cause of the issue.

Anthony
2 posts Page 1 of 1

Who is online

In total there are 19 users online :: 3 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot], Semrush [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry