Today I developed a method for toluene determination in my drug powder.
The instrument parameter are:
Head space: Tequilibrium: 80°C, Tneedle: 85°C; Ttransferline: 120°C. tequilibirum: 15 min. tinjection:0.03 min. Static head space mode.
GC: Colum zebron 624 30m x 0,32 mm x 1,80 um. Toven: 70 °C x 5 min increase 20°C min to 150°C. Tinjector: 210°C. Tdetector: 220°C. Carrier: 21PSI. Split: 5:1.
Standard preparatiron: Weight toluene in flask with DSMO, diluite it in other flask with water to the right concentration.
Vial of standard solution: 5 ml of STD solution
Sample vial: weight powder directly in vial and add the same quantitative of DMSO is present in STD solution and diluite to 5ml with water.
I motivate my choice:
- Diluite with water and not with DMSO because toluene is very poorly soluble in water and, in this mode, the excration by head space system is plus simply and effective. (with dilution in only DMSO the peak respound is little).
- 5 ml of solution in vial because K costant of toluene is very low. Then I work with high sample volume.
However I want ask you some things:
- Toluene boiling point is 110°C. If I work at 80°C in head space oven can I excract all of it from my powder? For this compound I know that the peak area respound dosen't change highly with temperature increase.The best choice for this analyte is to work with high liquid phase in HS vials.
- I discussed with my QA. I support that in the calculation for the standard concentration the volume of STD solution that I put into head space vials dosen't count. If I know that in my powder, for example, I weight 100mg of it, and that the limit in ppm of my residual solvent is 1000ppm I know that the final concentration of my standard solution is 0,1 mg/ml. This value is indipendent from volume in head space vials. If I put 100 ul or 10 ml in head space vial I've the same concentration. The only that change is component respound. Directly dipendent from his K value.
- My QA support, instead, that you must calculate, in head space analysis, standard concentration from the volume that you put into head space vials. Or if I put 1 ml of STD solution I must work with 0,1 mg/ml concentration. If I put 0,5 ml I must work with 0,2 mg/ml concentration ecc ecc
Who is right in your opinion?
If you have same idea to improve my method efficency, let me know please

Thank you very much!