Before this problem, I've been able to generate a pretty decent chromatogram for our analytes (B-carotene and other hydrophobic Vitamin A precursors). So I know what retention times I should be expecting for my samples when using the following mobile phases with a protocol gradient (I'm using a protocol that works fine at another university). Mobile A: 100% EtAc, Mobile B: 85%ACN:15%H2O:.1%TEA.
Everything seemed to be going fine except that after 1-2 weeks in which I was not working with the HPLC, I had some trouble with two overlapping peaks that have a similar retention time, and eluted as one fat peak. I tried a 50MeOH:50H2O:.1formic acid flush to clean up those particular peaks (which worked before to separate them), but this didn't work. I thought there might be contamination in the column so I flushed, & reverse flushed the column with isopropanol (I have a C18 column). When this didn't help, I made a 1M HNO3 solution thinking that maybe the detector needed to be cleaned. I removed the column and flushed overnight (....maybe not the best idea) with this 1M solution. (before the HNO3 solution, I did flush 100% ACN and then removed the column).
Since the HNO3 flush I haven't had any peaks from any samples or standards. I get the typical noise in the first minute as I have always gotten, which I assume comes from the changes in mobile phase, but there is no peaks past 1 minute even when I run for 40 minutes (usually the last compound elutes ~10min).
I've tried running 3 different columns all of which worked fine before and I get the same chromatographs (ie: I'm not getting any compound peaks on any of these columns), so I don't think there should be any problem with the column (especially since I didn't flush any nitric acid on any of the columns).
I extracted new samples to make sure that my samples hadn't deteriorated.
It's been almost a week since I ran the HNO3 so the whole system should have flushed out the acid by now. I checked the pH of mobile B before and after entering the line and it did drop .2, which would change the retention time, but not so much as to not get any peaks.
I suspect that my compounds may be eluting in the first minute: so how do I separate them back out?
or somehow the autosampler is not injecting the sample into the line? There is no leak anywhere and I can see the autosampler picking up the correctly programed sample into the port.
Help!
Thank you thank you thank you,
Tirzah