by
zheyin » Mon May 14, 2012 3:56 pm
wider slit = more light = less noise
wider slit = wider bandpass = worse spectral resolution
Since the goal is quantitation, not getting a UV spectrum, and since UV spectra in solution are fairly featureless, there is generally no reason to prefer a narrower slit. Assuming your wavelength is set to an absorbance maximum, at some point increasing the bandpass will let in more off-maximum light and the signal will start to decrease. That's the point at which you want to stop! So, to the extent that there is a guideline, it is "use the slit width that gives the best S/N ratio".
The response time has to stop at 10x higher, because higher response time will exceed the peak width of the major peaks and data points become not sufficient based on the guidelines from ChemStation.
That is why I also widened the slit from 4 nm to 8 nm.
But now I encounted another issue, after changing the slit several times, my chromatograms showed some wavy patterns,
it could show up in the middle of the run.
I doubt it is the detector malfunction.