Column Selection....

Posted: Tue May 08, 2012 11:38 am
by adamsmith87
I currently use an Agilent HP-ULTRA 2 to quantify Polyethylene Glycol in Carbowax.

I'm working on a method that will look at quantifying Propylene Glycol in a HydroGel. I've been getting good results from the PoraPLOT-Q column.

To avoid disturbing the system routinely, I was wondering if there is a suitable column I could use for both tests?

Regards.

Re: Column Selection....

Posted: Tue May 08, 2012 1:04 pm
by chromatographer1
IF
you can use the Agilent HP-ULTRA 2 for the propylene glycol
then you have the single column you desire.

I doubt there is any other possibility.

best wishes,

Rod

Re: Column Selection....

Posted: Tue May 08, 2012 1:15 pm
by adamsmith87
IF
you can use the Agilent HP-ULTRA 2 for the propylene glycol
then you have the single column you desire.

I doubt there is any other possibility.

best wishes,

Rod
Initial trials showed that the HP-ULTRA 2 didn't work with the PEG.

Based on that and what you say; it looks like we have no choice.

Thank you.