From what I gather, you're looking either for better resolution (separation between peaks) or greater sensitivity (greater height on peaks). If the former, then there could be two ways to approach it, depending on details you might need to give us:
1) You have the wrong type of column installed
2) You need to work on the acquisition method (temp. program, pneumatics, etc)
If the latter, you could approach it a few ways:
1) More concentrated sample
2) Lower split ratio
3) Splitless injection
Adding more of an analyte you're looking for is going to give you fabricated results--you'll end up seeing what you put in, right? There are times when you want to dope samples to see certain peaks change, but I'm guessing this isn't one of those times.
You need to give more details:
Column characteristics/type
Flows
Temp program
a sample chromatogram
Anything you can give to help us help you.
Best of luck!