Hello, I am trying to set up a method to analyze dichloroethane (dca) by Method 8011. I am trying to set up a oven parameters that help the dca elute after the hexane solvent peak but am having much difficulty. Hexane boils at 69 C, dca at 84 C and dibromoethane at 134 C so I'm thinking that it can be done I just need to find the right parameters. I have the oven starting at 70 C and holding for 2 min in an attempt to remove then hexane before the dca elutes. However, what I'm seeing is just a big hump instead of a nice peak for dca. Any suggestions?
-
- Posts: 124
- Joined: Tue Aug 02, 2011 5:11 am
What column you use?
"If your experiment needs statistics, you ought to have done a better experiment." Rutherford
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
pacerlaser,
You don't really ever want to start a run at the boiling point of your solvent. You don't get re-condensation on the front of the column but you also don't really get a good boil of the solvent. I would suggest starting cooler, like 45 or 50 to re-focus the solvent on the head of the column.
Best regards,
AICMM
You don't really ever want to start a run at the boiling point of your solvent. You don't get re-condensation on the front of the column but you also don't really get a good boil of the solvent. I would suggest starting cooler, like 45 or 50 to re-focus the solvent on the head of the column.
Best regards,
AICMM
-
- Posts: 145
- Joined: Sat Feb 27, 2010 5:27 am
The column is a ZB-624. I've tried using several different solvents now including pentane, hexane, methanol, and iso octane but still no luck. Thanks for the input so far!
-
- Posts: 1
- Joined: Fri May 04, 2012 2:58 am
As stated above, try starting at a lower temperature. Using methanol on that 624 column should allow for easy separation between the solvent and dca.
-
- Posts: 124
- Joined: Tue Aug 02, 2011 5:11 am
Zebron are good columns. Try to go lower in starting T, eg 60, or even 45. If you will use a thicker film, say 3 um you should be able to have starting T higher than for 0.1um. So it is a balance of resolution and speed.
Stay within the optimal liner velocity too.
Stay within the optimal liner velocity too.
"If your experiment needs statistics, you ought to have done a better experiment." Rutherford
-
- Posts: 350
- Joined: Tue Jun 28, 2011 7:30 pm
We use DB and RTX 624 columns; Zebron's is likely the same.
On a 30m x 0.25mm ID x 1.4 um film, helium at 1 mL/min, split 10:1. Start at 40 deg C, hold 1 min, 15 deg/min to 220, MS detector.
Direct injection of a VOA mix in MeOH gives us MeOH ~2 to 3 min (we do not scan that low so it just a baseline disturbance). The 1,2-DCA peak comes out at 5.8 minutes (coelutes with Benzene).
IIRC, 8011 uses an ELCD detector; if so the solvent should be vented or it will char up the nickel catalyst. A GC detector at atmosopheric pressure will give some different behavior than an MSD at vacuum, but you column should be able to separate the DCA (likely better than the MSD set up).
Hexane may well elute too close, I can't recall where that comes out related to DCA.
Tim
On a 30m x 0.25mm ID x 1.4 um film, helium at 1 mL/min, split 10:1. Start at 40 deg C, hold 1 min, 15 deg/min to 220, MS detector.
Direct injection of a VOA mix in MeOH gives us MeOH ~2 to 3 min (we do not scan that low so it just a baseline disturbance). The 1,2-DCA peak comes out at 5.8 minutes (coelutes with Benzene).
IIRC, 8011 uses an ELCD detector; if so the solvent should be vented or it will char up the nickel catalyst. A GC detector at atmosopheric pressure will give some different behavior than an MSD at vacuum, but you column should be able to separate the DCA (likely better than the MSD set up).
Hexane may well elute too close, I can't recall where that comes out related to DCA.
Tim
7 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science