The problem: If we lower the split ratio below 1:20 the internal pressure drops slowly until a "CAR1: Purge Leaks" error is displayed. We used to run 1:5 until this started, at which point 1:10 worked, then only 1:15 worked, now anything below 1:20 will not work. We are using glass inserts for split use with glass wool (we put the wool in ourselves), and an HP-chiral 20b column (30 m, 0.32-mm internal diameter, 0.25-mm film thickness; Agilent Technologies), He carrier gas.
Question: Assuming no real leaks, is this a problem in the Math that the TFC and APC are calculating from the sensors or settings, or is this a broken TFC resistor, valve, etc. Is it possible to have a bad setting somewhere?
What I have done: Snooped all lines and found no leaks. I followed the manual protocol for detecting leaks:
(In short) Set Flow control to 'Continuous', Turn off, All temps < 40 C,
Remove column and seal injector and detector with graphite ferrule. Blank nut Purge and Split outlets.
Cylinder pressure at 400 kPa,
Set Split mode to 'Direct', Control mode to 'Pressure'.
Set Inlet Pressure to 150 kPa,
Set Purge flow to 500 ml/min,
Press [On/Off] to start AFC control.
Result: Inlet pressure stays at 150 kPa. Total Flow is 20 ml/min (Indicates Leak). Snooped all connections no leaks detected, including injection port connections, septum, O-ring (flouride). (We don't have an electronic detector, I do my best to prevent drips etc).
When I Turn the system off again: Inlet pressure climbs very slowly (1kPa/5min) It never decreases (Indicating no negative leak). Total flow remains constant at 20 ml/min (Why? How?).
When I release the pressure: Inlet pressure drops to 0 [System is open at the column nut], Total flow remains at 20 ml/min. Is the TFC control valve leaking? Or is it reporting a confused number?
When External pressure is increased or decreased, Total Flow at zero inlet pressure changes roughly according to the following:
{Inlet Pressure}, External Pressure, [Total Flow]
{0 ml/min}, 900 kPa, [30 ml/min]
{0 ml/min}, 400 kPa, [20 ml/min]
{0 ml/min}, 180 kPa, [10 ml/min]
(Is external pressure leaking into the system dependent on it's pressure?)
I also removed the injection port lower conections to check for broken glass, cleaned and reinstalled: No leaks before or after.
So does this sound like broken parts or a problem with the way we are using the machine? Thank you for your help!
John