Posted: Fri May 04, 2012 10:16 am
Hi all, apologies if this question is simple but my experience is limited.
We are using a Thermo HPLC system attached to an LCQ MS/MS with a 2.1mm C18 column. We are running gradients of ACN (+0.01% formic acid) and water (+0.01% formic acid) with a flow rate of 0.3 ml/min. We are injecting 10ul of sample in 66% ACN.
We are trying to analyse rhodamine 123 via MS/MS (we have a method for HPLC fluorsecence analysis using a 4.6mm column that works fine with this molecule). The gradient we use is the same as the one for fluorescence, injection at 5% ACN then increase to 95% and hold for a few minutes before equilibrating for next injection. However we are having issues the compound retention on the column. Of a standard curve run 2 peaks came out with a retention time of 12 minutes as expected and 7 came out in the solvent front, there seemed to be no correlation with concentration.
We are having similar problems with loepramide and I am starting to worry about the condition of the column although it is new.
We have tried -
- methanol instead of ACN
- samller injection volumes
- injecting in just water ratehr than 66% ACN
This has really stumped me and I am grateful for ay advice on how to proceed and sort this out.
Dan
We are using a Thermo HPLC system attached to an LCQ MS/MS with a 2.1mm C18 column. We are running gradients of ACN (+0.01% formic acid) and water (+0.01% formic acid) with a flow rate of 0.3 ml/min. We are injecting 10ul of sample in 66% ACN.
We are trying to analyse rhodamine 123 via MS/MS (we have a method for HPLC fluorsecence analysis using a 4.6mm column that works fine with this molecule). The gradient we use is the same as the one for fluorescence, injection at 5% ACN then increase to 95% and hold for a few minutes before equilibrating for next injection. However we are having issues the compound retention on the column. Of a standard curve run 2 peaks came out with a retention time of 12 minutes as expected and 7 came out in the solvent front, there seemed to be no correlation with concentration.
We are having similar problems with loepramide and I am starting to worry about the condition of the column although it is new.
We have tried -
- methanol instead of ACN
- samller injection volumes
- injecting in just water ratehr than 66% ACN
This has really stumped me and I am grateful for ay advice on how to proceed and sort this out.
Dan