The GPV test passed the 5% criteria, but was not ideal (1% or less) at 2.29%. The plateau for the A-D valve combination was a little taller than those for the other 3 combinations, which agreed very well between themselves, so I suspected valve D might be a little off.
There were step tests for A-C, B-D, A-D, and B-C combinations. The article said to compare the step test to a theoretical programmed value. How do you get a theoretical value without running an acetone calibration? I predicted theoretical values by measuring the height of the first (10%) step and then assuming the 2nd (20%) step would be twice as tall, etc. Is this what was meant? I also tried calculating an average corrected delta step size and using that for the theoretical value.
Most of the steps agreed more or less within the 1% criteria using the former approach, but error at the highest step (100% 0.1% acetone solution) was always a bit higher whenever valve D was used. I suspected valve D already, so that made some sense. However, for all of the combinations, once I got above about 50% of the 0.1% acetone solution, delta step size decreased as proportion of the acetone solution increased.
Linearity tests for A-C and B-D combinations were run, and also one for A+B and C+D (i.e. a 50:50:0:0 to 0:0:50:50 gradient). These results looked nearly perfect.
I don’t understand how the linear gradients could appear so perfect when the step increments were shrinking as the UV absorber concentration increased. Any insights?