I am running Agilent 1100 for impurities in mannitol (sugar). Column temp: 85 C, Flow Rate: 0.5 ml/min, Column: Carbohydrate column, Injection Volume: 20 uL, Mobile Phase: DI Water, Diluent: DI Water, Detector: Refractive Index.
I see a negative peak at around 17 mins. It also occurs in blank.
When I reduced the injection volume of my blank from 20 uL to 10 uL the area of this peak reduced by almost half.
When I reduced the flow rate by two times, the peak Retention Time increased by two times.
When I ran DI Water from a different source, same peak appears.
It is a very small negative peak but I want to justify its existence.
Any comments of what could it be? Impurity in water? Something sticking to my injector loop? or something inherent to RI detector?
Thanks!!