how do I choose syringe filter in order to minimize adsorption of hydrophobic analytes in aqueous phase? nylon? PTFE? GSH?
-
- Posts: 621
- Joined: Tue Nov 27, 2007 1:20 pm
For Aqueous Phases PTFE or PP filters are not normally recommended.
you may have to run some comparative work with other types listed in the selection guides http://www.chromtech.com/Catalog7/PDF_Indvpge/284.pdf
you may have to run some comparative work with other types listed in the selection guides http://www.chromtech.com/Catalog7/PDF_Indvpge/284.pdf
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
-
- Posts: 124
- Joined: Tue Aug 02, 2011 5:11 am
There are recommendations, but for the final tuning you may need to do experiments.
Also watch for leaching interferences. Ultimately – centrifuge.
Also watch for leaching interferences. Ultimately – centrifuge.
"If your experiment needs statistics, you ought to have done a better experiment." Rutherford
-
- Posts: 206
- Joined: Sun Oct 09, 2011 5:08 pm
this selection guide is for protein. do you have selection guide based on small molecules.
Anyway, hydrophobic interaction is responsible for protein binding, right?
Anyway, hydrophobic interaction is responsible for protein binding, right?
For Aqueous Phases PTFE or PP filters are not normally recommended.
you may have to run some comparative work with other types listed in the selection guides http://www.chromtech.com/Catalog7/PDF_Indvpge/284.pdf
-
- Posts: 526
- Joined: Fri Feb 10, 2012 7:39 pm
For aqueous samples we typically use nylon or glass membrane filters, such as GE Whatman Autovials. But we do check for recovery/interferences.
For organic or mixed organic-aqueous, we typically use PTFE.
For organic or mixed organic-aqueous, we typically use PTFE.
-
- Posts: 658
- Joined: Thu Jun 19, 2008 11:07 am
In worst case you may also have to consider the syringe (plastic-glass), not often but happned to us that plastic disposeble syringe adsorbs the analyte.
Izaak Kolthoff: “Theory guides, experiment decides.”
6 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 0 registered, 0 hidden and 2 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 2 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 2 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science