Karen01 » Tue Apr 24, 2012 6:46 pm
how can carrier gas going out throrugh the septum purge dilute the sample ?. Even if you have the headspacer plumbed to supply all the gas to the inlet, any of the sample going out to the septum purge simply increases the effective split ratio.
That is what I meant.
The 1 ml default sample loops of valve and loop samplers are a hold over from the days of packed columns. To work effectively with capillary columns they need split injections, and that means having an inlet connection with extra gas going in and out, and issues of poorly swept volumes, adsorption and gas mixing. To be compatible with capillary GC the samples need loop volumes of 20 - 100 ul, with a zero dead volume connection between the transfer line and the column (or better yet with the column connected direct to the six-port valve).
For the loop sampler I always used split injections
With either loop or syringe injections and with splitless transfer to columns up to and including 0.32 mm diameter through a conventional split-splitless inlet, the gas flow rate through the inlet is too slow to effectively flush the internal volume quickly enough to ensure narrow peaks for compounds that are not stationary phase focussed - especially if the inlet has a "splitless" liner which has a larger internal volume in order to accommodate the vapour cloud from a hot injection of volatile solvent.
I use 2 mm straight through liner for headspace.
With syringe injections there is the additional problem that the inlet pressure causes back-flow into the syringe as the needle penetrates the septum, with subsequent issues of incomplete mixing and incomplete transfer of the original sample volume into the inlet. The position of the tip of the needle or transfer line relative to the top of the column can impact what fraction of the headspace sample gets into the column, and what fraction flows away down the split outlet or diffuses into contact with active metal surfaces.[/color]

I would think that this is a typo, except that 20 ml/min is still too fast, and 20 ul/s is too slow, giving a minimum starting band width of 17.5 s - so maybe the devil is in the details[/color]
It should have been 20 uL/sec. The advice I've seen here for syringe headspace samplers is to use half the carrier flow rate. Half of 2.5 mL/min = 20.8 ul/sec and that is what I used...
Injecting larger volumes using a split injection to get faster flow does not get me much from what I can see. Is that wrong?
Thanks,
- Karen